Dictionary: cal·i·bra·tion   (kăl'ə-brā'shən) pronunciation
n.
  1. The act or process of calibrating or the state of being calibrated.
  2. A set of gradations that show positions or values. Often used in the plural: the calibrations on a pressure gauge.

Search unanswered questions...
Enter a question here...
Search: All sources Community Q&A Reference topics
Sci-Tech Encyclopedia: Calibration
Top

The process of determining the performance parameters of an artifact, instrument, or system by comparing it with measurement standards. Adjustment may be a part of a calibration, but not necessarily. A calibration assures that a device or system will produce results which meet or exceed some defined criteria with a specified degree of confidence.

Two important measurement concepts related to calibration are precision and accuracy. Precision refers to the minimum discernible change in the parameter being measured, while accuracy refers to the actual amount of error that exists in a calibration. All measurement processes used for calibration are subject to various sources of error. It is common practice to classify them as random or systematic errors. When a measurement is repeated many times, the results will exhibit random statistical fluctuations which may or may not be significant. Systematic errors are offsets from the true value of a parameter and, if they are known, corrections are generally applied, eliminating their effect on the calibration. If they are not known, they can have an adverse effect on the accuracy of the calibration. High-accuracy calibrations are usually accompanied by an analysis of the sources of error and a statement of the uncertainty of the calibration. Uncertainty indicates how much the accuracy of a calibration could be degraded as a result of the combined errors. See also Analysis of variance; Distribution (probability); Instrument science; Physical measurement; Probability; Statistics.


Dental Dictionary: calibration
Top

n

The process of comparing a measurement instrument against a verified standard instrument. The US Bureau of Standards maintains the national calibration instruments for weights and measures.

Geography Dictionary: calibration
Top

The adjustment of a model so that it will fit special circumstances. The values used within a particular model, such as the gravity model, need to be modified to local circumstances before it can make predictions. Thus, complex versions of the gravity model build in the variables of job opportunities, wage rates, and prices, but the amount of weight given to each of these variables will vary from time to time and from place to place. Calibration is a method of adjusting their weighting.

Archaeology Dictionary: calibration
Top

(calibrated radiocarbon age/ calibrated date) [De]

When using radiocarbon dating relating to archaeological materials from the last 10 000 years or so it is necessary to standardize or calibrate the laboratory determination (conventional radiocarbon age) to convert it into calendar years. This is especially important if radiocarbon determinations are to be compared with historically derived calendar dates, or dates determined by other means.

Laboratory determinations are given in radiocarbon years before present (RCYBP or BP), but radiocarbon years are not of equal length because of variations in the level of 14C in the atmosphere in the past. Special corrections for the effects of certain factors that alter the background level of 14C may also be necessary. Calibration is based on measurements of the 14C levels in material of known age, principally samples of ancient wood taken from dendrochronology dated sequences. The earliest calibration curves, published in the mid 1970s, used samples from the bristlecone pine (Pinus aristat), an especially long-lived tree species that grows in the Sierra Nevada of California. During the 1980s and 1990s regional calibration curves were developed for different parts of the world based on local dendrochronological sequences. A number of excellent computer-based calibration programmes are available, including OxCal from the Radiocarbon Accelerator Unit in Oxford University, UK, and CALIB from the University of Washington in Seattle, USA. Both can be downloaded from the world wide web.

Calibrated radiocarbon ages are conventionally indicated by the abbreviations bc/ad or cal.bc/cal.ad. However, because the calibration process involves using the standard deviations provided by the laboratory for the original determination, the calibration of a specific determination will result in an age range, the width of which will depend on the size of the standard deviation, the level of certainty used (typically 66 per cent or 95 per cent based on one and two standard deviations respectively), and the nature of the calibration curve in the region covered by the initial determination. Thus, calibrated radiocarbon dates are usually given as a range, for example 2030–2125 bc, which if calibrated using a single place of standard deviation would mean that there was a 66 per cent probability that the actual date lies within that range. When citing dates it is standard practice to give the calibrated age, the original determination, and the laboratory code (e.g. 2472–2404 bc (3925 ± 35 RCYBP) OxA-5328). Within a book or academic paper all dates should be calibrated using the same procedures and the same calibration curve, and the parameters used should be summarized in the preface or some other convenient place. See also radiocarbon dating, conventional radiocarbon age, and corrected age.

The determination of the accuracy of an instrument by measuring its variation from a known standard.

Veterinary Dictionary: calibration
Top

Determination of the accuracy of an instrument, usually by measurement of its variation from a standard, to ascertain necessary correction factors.

Electronics Dictionary: calibration
Top

To adjust the correct value of a reading by comparison to a standard.


Wikipedia: Calibration
Top

Calibration is the set of operations that establish, under specified conditions, the relationship between the values of quantities indicated by a measuring instrument and the corresponding values realized by standards. International Vocabulary of Metrology (VIM)

The result of a calibration permits either the assignment of values of measurands to the indications or the determination of corrections with respect to indications.

A calibration may also determine other metrological properties such as the effect of influence quantities.

The result of a calibration may be recorded in a document, sometimes called a calibration certificate or a calibration report.

A measuring instrument can be calibrated by comparison with a standard.

An adjustment of the instrument is often carried out after calibration in order that it provides given indications corresponding to given values of the quantity measured.

When the instrument is made to give a null indication corresponding to a null value of the quantity to be measured, the set of operation is called zero adjustment .


Contents

History

Many of the earliest measuring devices were intuitive and easy to conceptually validate. The term "calibration" probably was first associated with the precise division of linear distance and angles using a dividing engine and the measurement of gravitational mass using a weighing scale. These two forms of measurement alone and their direct derivatives supported nearly all commerce and technology development from the earliest civilizations until about 1800AD.

The Industrial Revolution introduced wide scale use of indirect measurement. The measurement of pressure was an early example of how indirect measurement was added to the existing direct measurement of the same phenomena.

Direct reading design
Indirect reading design from front
Indirect reading design from rear, showing Bourdon tube

Before the Industrial Revolution, the most common pressure measurement device was a hydrostatic manometer, which is not practical for measuring high pressure. Eugene Bourdon filled the need for high pressure measurement with his Bourdon tube pressure gage.

In the direct reading hydrostatic manometer design on the left, unknown pressure pushes the liquid down the left side of the manometer U-tube (or unknown vacuum pulls the liquid up the tube, as shown) where a length scale next to the tube measures the pressure, referenced to the other, open end of the manometer on the right side of the U-tube. The resulting height difference "H" is a direct measurement of the pressure or vacuum with respect to atmospheric pressure. The absence of pressure or vacuum would make H=0. The self-applied calibration would only require the length scale to be set to zero at that same point.

In a Bourdon tube shown in the two views on the right, applied pressure entering from the bottom on the silver barbed pipe tries to straighten a curved tube (or vacuum tries to curl the tube to a greater extent), moving the free end of the tube that is mechanically connected to the pointer. This is indirect measurement that depends on calibration to read pressure or vacuum correctly. No self-calibration is possible, but generally the zero pressure state is correctable by the user.

Even in recent times, direct measurement is used to increase confidence in the validity of the measurements.

[[Image:

In the early days of US automobile use, people wanted to see the gasoline they were about to buy in a big glass pitcher, a direct measure of volume and quality via appearance. By 1930, rotary flowmeters were accepted as indirect substitutes. Visible in the picture above on the right (above the price digit register) is a hemispheric viewing window where consumers could see the blade of the flowmeter turn as the gasoline was pumped. By 1970, the windows were gone and the measurement was totally indirect.

Indirect measurement always involve linkages or conversions of some kind. It is seldom possible to intuitively monitor the measurement. These facts intensify the need for calibration.

Most measurement techniques used today are indirect.

Calibration versus Metrology

There is no consistent demarcation between calibration and metrology. Generally, the basic process below would be metrology-centered if it involved new or unfamiliar equipment and processes. For example, a calibration laboratory owned by a successful maker of microphones would have to be proficient in electronic distortion and sound pressure measurement. For them, the calibration of a new frequency spectrum analyzer is a routine matter with extensive precedent. On the other hand, a similar laboratory supporting a coaxial cable manufacturer may not be as familiar with this specific calibration subject. A transplanted calibration process that worked well to support the microphone application may or may not be the best answer or even adequate for the coaxial cable application. A prior understanding the measurement requirements of coaxial cable manufacturing would make the calibration process below more successful.

Basic calibration process

The calibration process begins with the design of the measuring instrument that needs to be calibrated. The design has to be able to "hold a calibration" through its calibration interval. In other words, the design has to be capable of measurements that are "within engineering tolerance" when used within the stated environmental conditions over some reasonable period of time. Having a design with these characteristics increases the likelihood of the actual measuring instruments performing as expected.

The exact mechanism for assigning tolerance values varies by country and industry type. The measuring equipment manufacturer generally assigns the measurement tolerance, suggests a calibration interval and specifies the environmental range of use and storage. The using organization generally assigns the actual calibration interval, which is dependent on this specific measuring equipment's likely usage level. A very common interval in the United States for 8–12 hours of use 5 days per week is six months. That same instrument in 24/7 usage would generally get a shorter interval. The assignment of calibration intervals can be a formal process based on the results of previous calibrations.

The next step is defining the calibration process. The selection of a standard or standards is the most visible part of the calibration process. Ideally, the standard has less than 1/4 of the measurement uncertainty of the device being calibrated. When this goal is met, the accumulated measurement uncertainty of all of the standards involved is considered to be insignificant when the final measurement is also made with the 4:1 ratio. This ratio was probably first formalized in Handbook 52 that accompanied MIL-STD-45662A, an early US Department of Defense metrology program specification. It was 10:1 from its inception in the 1950s until the 1970s, when advancing technology made 10:1 impossible for most electronic measurements.

Maintaining a 4:1 accuracy ratio with modern equipment is difficult. The test equipment being calibrated can be just as accurate as the working standard. If the accuracy ratio is less than 4:1, then the calibration tolerance can be reduced to compensate. When 1:1 is reached, only an exact match between the standard and the device being calibrated is a completely correct calibration. Another common method for dealing with this capability mismatch is to reduce the accuracy of the device being calibrated.

For example, a gage with 3% manufacturer-stated accuracy can be changed to 4% so that a 1% accuracy standard can be used at 4:1. If the gage is used in an application requiring 16% accuracy, having the gage accuracy reduced to 4% will not affect the accuracy of the final measurements. This is called a limited calibration. But if the final measurement requires 10% accuracy, then the 3% gage never can be better than 3.3:1. Then perhaps adjusting the calibration tolerance for the gage would be a better solution. If the calibration is performed at 100 units, the 1% standard would actually be anywhere between 99 and 101 units. The acceptable values of calibrations where the test equipment is at the 4:1 ratio would be 96 to 104 units, inclusive. Changing the acceptable range to 97 to 103 units would remove the potential contribution of all of the standards and preserve a 3.3:1 ratio. Continuing, a further change to the acceptable range to 98 to 102 restores more than a 4:1 final ratio.

This is a simplified example. The mathematics of the example can be challenged. It is important that whatever thinking guided this process in an actual calibration be recorded and accessible. Informality contributes to tolerance stacks and other difficult to diagnose post calibration problems.

Also in the example above, ideally the calibration value of 100 units would be the best point in the gage's range to perform a single-point calibration. It may be the manufacturer's recommendation or it may be the way similar devices are already being calibrated. Multiple point calibrations are also used. Depending on the device, a zero unit state, the absence of the phenomenon being measured, may also be a calibration point. Or zero may be resettable by the user-there are several variations possible. Again, the points to use during calibration should be recorded.

There may be specific connection techniques between the standard and the device being calibrated that may influence the calibration. For example, in electronic calibrations involving analog phenomena, the impedance of the cable connections can directly influence the result.

All of the information above is collected in a calibration procedure, which is a specific test method. These procedures capture all of the steps needed to perform a successful calibration. The manufacturer may provide one or the organization may prepare one that also captures all of the organization's other requirements. There are clearinghouses for calibration procedures such as the Government-Industry Data Exchange Program (GIDEP) in the United States.

This exact process is repeated for each of the standards used until transfer standards, certified reference materials and/or natural physical constants, the measurement standards with the least uncertainty in the laboratory, are reached. This establishes the traceability of the calibration.

See metrology for other factors that are considered during calibration process development.

After all of this, individual instruments of the specific type discussed above can finally be calibrated. The process generally begins with a basic damage check. Some organizations such as nuclear power plants collect "as-found" calibration data before any routine maintenance is performed. After routine maintenance and deficiencies detected during calibration are addressed, an "as-left" calibration is performed.

More commonly, a calibration technician is entrusted with the entire process and signs the calibration certificate, which documents the completion of a successful calibration.

Calibration process success factors

The basic process outlined above is a difficult and expensive challenge. The cost for ordinary equipment support is generally about 10% of the original purchase price on a yearly basis, as a commonly accepted rule-of-thumb. Exotic devices such as scanning electron microscopes, gas chromatograph systems and laser interferometer devices can be even more costly to maintain.

The extent of the calibration program exposes the core beliefs of the organization involved. The integrity of organization-wide calibration is easily compromised. Once this happens, the links between scientific theory, engineering practice and mass production that measurement provides can be missing from the start on new work or eventually lost on old work.

The 'single measurement' device used in the basic calibration process description above does exist. But, depending on the organization, the majority of the devices that need calibration can have several ranges and many functionalities in a single instrument. A good example is a common modern oscilloscope. There easily could be 200,000 combinations of settings to completely calibrate and limitations on how much of an all inclusive calibration can be automated.

Every organization using oscilloscopes has a wide variety of calibration approaches open to them. If a quality assurance program is in force, customers and program compliance efforts can also directly influence the calibration approach. Most oscilloscopes are capital assets that increase the value of the organization, in addition to the value of the measurements they make. The individual oscilloscopes are subject to depreciation for tax purposes over 3, 5, 10 years or some other period in countries with complex tax codes. The tax treatment of maintenance activity on those assets can bias calibration decisions.

New oscilloscopes are supported by their manufacturers for at least five years, in general. The manufacturers can provide calibration services directly or through agents entrusted with the details of the calibration and adjustment processes.

Very few organizations have only one oscilloscope. Generally, they are either absent or present in large groups. Older devices can be reserved for less demanding uses and get a limited calibration or no calibration at all. In production applications, oscilloscopes can be put in racks used only for one specific purpose. The calibration of that specific scope only has to address that purpose.

This whole process in repeated for each of the basic instrument types present in the organization, such as the digital multimeter (DMM) pictured below.

A DMM (top), a rack-mounted oscilloscope (center) and control panel

Also the picture above shows the extent of the integration between Quality Assurance and calibration. The small horizontal unbroken paper seals connecting each instrument to the rack prove that the instrument has not been removed since it was last calibrated. These seals are also used to prevent undetected access to the adjustments of the instrument. There also are labels showing the date of the last calibration and when the calibration interval dictates when the next one is needed. Some organizations also assign unique identification to each instrument to standardize the recordkeeping and keep track of accessories that are integral to a specific calibration condition.

When the instruments being calibrated are integrated with computers, the integrated computer programs and any calibration corrections are also under control.

In the United States, there is no universally accepted nomenclature to identify individual instruments. Besides having multiple names for the same device type there also are multiple, different devices with the same name. This is before slang and shorthand further confuse the situation, which reflects the ongoing open and intense competition that has prevailed since the Industrial Revolution.

The calibration paradox

Successful calibration has to be consistent and systematic. At the same time, the complexity of some instruments require that only key functions be identified and calibrated. Under those conditions, a degree of randomness is needed to find unexpected deficiencies. Even the most routine calibration requires a willingness to investigate any unexpected observation.

Theoretically, anyone who can read and follow the directions of a calibration procedure can perform the work. It is recognizing and dealing with the exceptions that is the most challenging aspect of the work. This is where experience and judgement are called for and where most of the resources are consumed.

Additional information

Calibration is the process of establishing the relationship between a measuring device and the units of measure. This is done by comparing a device or the output of an instrument to a standard having known measurement characteristics. For example the length of a stick can be calibrated by comparing it to a standard that has a known length. Once the relationship of the stick to the standard is known the stick is calibrated and can be used to measure the length of other things.

For many operations the quality of the calibration needs to be known and is quantified by an uncertainty estimate for the calibration. This is so important for the scientific community and manufacturing operations that it has been proposed that an evaluation of the measurement uncertainty was added as part of the calibration process.[1]

Part of calibration is to zero the measuring device, the process of establishing that the zero point of the device corresponds to zero on the relevant scale.

Quality

To improve the quality of the calibration and have the results accepted by outside organizations it is desirable for the calibration and subsequent measurements to be "traceable" to the internationally defined measurement units. Establishing traceability is accomplished by a formal comparison to a standard which is directly or indirectly related to national standards (NIST in the USA), international standards, or certified reference materials.

Quality management systems call for an effective metrology system which includes formal, periodic, and documented calibration of all measuring instruments. ISO 9000 and ISO 17025 sets of standards require that these traceable actions are to a high level and set out how they can be quantified.

Instrument calibration

Calibration can be called for:

  • with a new instrument
  • when a specified time period is elapsed
  • when a specified usage (operating hours) has elapsed
  • when an instrument has had a shock or vibration which potentially may have put it out of calibration
  • whenever observations appear questionable

In non-specialized use, calibration is often regarded as including the process of adjusting the output or indication on a measurement instrument to agree with value of the applied standard, within a specified accuracy. For example, a thermometer could be calibrated so the error of indication or the correction is determined, and adjusted (e.g. via calibration constants) so that it shows the true temperature in Celsius at specific points on the scale.

There is good software available for doing this now, recording measurement results and producing Calibration Certificates are just some of the advantages. See the 'External links' below for more information.

International

In many countries a National Metrology Institute (NMI) will exist which will maintain primary standards of measurement (the main SI units plus a number of derived units) which will be used to provide traceability to customer's instruments by calibration. The NMI supports the metrological infrastructure in that country (and often others) by establishing an unbroken chain, from the top level of standards to an instrument used for measurement. Examples of National Metrology Institutes are NPL in the UK, NIST in the United States, PTB in Germany and many others. Since the Mutual Recognition Agreement was signed it is now straightforward to take traceability from any participating NMI and it is no longer necessary for a company to obtain traceability for measurements from the NMI of the country in which it is situated.

See [1] for more information on the Mutual Recognition Agreement. See [2] for the international metrology framework.

To communicate the quality of a calibration the calibration value is often accompanied by a traceable uncertainty statement to a stated confidence level. This is evaluated through careful uncertainty analysis.

See also

References

  1. ^ The International Vocabulary of Basic and General Terms in Metrology (VIM) draft April 2004

External links

Further reading

  • Pyzdek, T, "Quality Engineering Handbook", 2003, ISBN 0824746147
  • Godfrey, A. B., "Juran's Quality Handbook", 1999, ISBN 007034003

Translations: Calibration
Top

Dansk (Danish)
n. - kalibrering, korrigering, korrektionsbestemmelse

Nederlands (Dutch)
schaalverdeling, ijking, kaliberbepaling

Français (French)
n. - calibrage, étalonnage, tarage

Deutsch (German)
n. - Eichung, Skala

Ελληνική (Greek)
n. - διαμέτρηση, διαβάθμιση, καλιμπράρισμα, διακρίβωση, βαθμονόμηση, μέτρηση της (εσωτερικής) διαμέτρου

Italiano (Italian)
calibratura, taratura

Português (Portuguese)
n. - calibragem (f), aferição (f)

Русский (Russian)
калибровка

Español (Spanish)
n. - escala graduada, calibración, ajuste

Svenska (Swedish)
n. - kalibrering, gradstreck

中文(简体)(Chinese (Simplified))
口径测定, 划刻度, 查刻度

中文(繁體)(Chinese (Traditional))
n. - 口徑測定, 劃刻度, 查刻度

한국어 (Korean)
n. - 구경 측정, 눈금

日本語 (Japanese)
n. - 口径測定, 目盛り定め, 較正

العربيه (Arabic)
‏(الاسم) تعيير, تدريج, تقويم‏

עברית (Hebrew)
n. - ‮כיול‬


Best of the Web: calibration
Top

Some good "calibration" pages on the web:


Math
mathworld.wolfram.com
 
Shopping: calibration
Top
 
 
Learn More
calibration point (ordnance)
check standard (analytical chemistry)
monitor calibrator (technology)

How do you calibrate for HPLC? Read answer...
How do you calibrate on a dsi? Read answer...
What does calibrated mean? Read answer...

Help us answer these
Why get your tv calibrated?
How do you calibrate inclinator?
Can you calibrate the thermometer?

Post a question - any question - to the WikiAnswers community:

 

Copyrights:

Dictionary. The American Heritage® Dictionary of the English Language, Fourth Edition Copyright © 2007, 2000 by Houghton Mifflin Company. Updated in 2009. Published by Houghton Mifflin Company. All rights reserved.  Read more
Sci-Tech Encyclopedia. McGraw-Hill Encyclopedia of Science and Technology. Copyright © 2005 by The McGraw-Hill Companies, Inc. All rights reserved.  Read more
Dental Dictionary. Mosby's Dental Dictionary. Copyright © 2004 by Elsevier, Inc. All rights reserved.  Read more
Geography Dictionary. A Dictionary of Geography. Copyright © Susan Mayhew 1992, 1997, 2004. All rights reserved.  Read more
Archaeology Dictionary. The Concise Oxford Dictionary of Archaeology. Copyright © 2002, 2003 by Oxford University Press. All rights reserved.  Read more
Sports Science and Medicine. The Oxford Dictionary of Sports Science & Medicine. Copyright © Michael Kent 1998, 2006, 2007. All rights reserved.  Read more
Veterinary Dictionary. Saunders Comprehensive Veterinary Dictionary 3rd Edition. Copyright © 2007 by D.C. Blood, V.P. Studdert and C.C. Gay, Elsevier. All rights reserved.  Read more
Electronics Dictionary. Copyright 2001 by Twysted Pair. All rights reserved.  Read more
Wikipedia. This article is licensed under the Creative Commons Attribution/Share-Alike License. It uses material from the Wikipedia article "Calibration" Read more
Translations. Copyright © 2007, WizCom Technologies Ltd. All rights reserved.  Read more