n.
A microscope that scans the surface of a sample with a beam of electrons, causing a narrow channel of tunneling electrons to flow between the sample and the beam, and producing three-dimensional images of atomic topography and structure.
| Dictionary: scanning tunneling microscope |
A microscope that scans the surface of a sample with a beam of electrons, causing a narrow channel of tunneling electrons to flow between the sample and the beam, and producing three-dimensional images of atomic topography and structure.
| Sci-Tech Encyclopedia: Scanning tunneling microscope |
An instrument for producing surface images with atomic-scale lateral resolution, in which a fine probe tip is scanned over the surface at a distance of 0.5–1 nanometer, and the resulting tunneling current, or the position of the tip required to maintain a constant tunneling current, is monitored.
Scanning tunneling microscopes have pointed electrodes that are scanned over the surface of a conducting specimen, with help from a piezoelectric crystal whose dimensions can be altered electronically. They normally generate images by holding the current between the tip of the electrode and the specimen at some constant (set-point) value by using a piezoelectric crystal to adjust the distance between the tip and the specimen surface, while the tip is piezoelectrically scanned in a raster pattern over the region of specimen surface being imaged. By holding the force, rather than the electric current, between tip and specimen at a set-point value, atomic force microscopes similarly allow the exploration of nonconducting specimens. In either case, when the height of the tip is plotted as a function of its lateral position over the specimen, an image that looks very much like the surface topography results.
It is becoming increasingly possible to record other signals (such as lateral force, capacitance, scan-related tip displacement, temperature, light intensity, or magnetic resonance) as the tip scans. For example, modern atomic force microscopes can map lateral force and conductivity along with height, while image pairs from scanning tunneling microscopes scanning to and fro can provide information about friction as well as topography.
Scanning tunneling microscopes make it possible not just to view atoms but to push them and even to rearrange them in unlikely combinations (sometimes whether or not these rearrangements are desirable). A few considerations of scale are important in understanding this process. Atoms comprise a positive nucleus and a surrounding cloud of negative electrons. These charges rearrange when another atom approaches, with unlike charges shifting to give rise to the van der Waals force of attraction between neutral atoms. This force makes gravity (and most accelerations) ignorable when contact between solid objects in the micrometer size range and smaller is involved, since surface-to-volume ratios are inversely proportional to object size.
The electric field in the scanning tunneling microscope allows plucking as well, in which adsorbed or substrate atoms are removed and transferred to the electrode tip with a suitable voltage pulse. Because the electric field from the tip falls off less rapidly with separation than do van der Waals forces, the most weakly attached nearby atom rather than the nearest may end up being removed. One solution to this problem is a hybrid approach. By invoking the tip electric field for bond breaking only when the tip is sufficiently close to the target atom that the van der Waals forces contribute as well, atoms on silicon could be singly removed and redeposited at will.
A third kind of selective bond breaking was also demonstrated. It involved the selective breaking of silicon-hydrogen bonds using electron energies (that is, pulse voltages) below those necessary to break bonds directly. Since the desorption probability was observed to vary exponentially with the tip-specimen current, it is believed that vibrational heating from inelastic electron tunneling mediated the chemical transition in this work. This work involves bond alteration at the level of signal atoms, the ultimate frontier for lithographic miniaturization.
| Columbia Encyclopedia: scanning tunneling microscope |
| Wikipedia: Scanning tunneling microscope |
Scanning tunneling microscopy (STM) is a powerful technique for viewing surfaces at the atomic level. Its development in 1981 earned its inventors, Gerd Binnig and Heinrich Rohrer (at IBM Zürich), the Nobel Prize in Physics in 1986[1][2]. STM probes the density of states of a material using tunneling current. For STM, good resolution is considered to be 0.1 nm lateral resolution and 0.01 nm depth resolution[3]. The STM can be used not only in ultra high vacuum but also in air and various other liquid or gas ambients, and at temperatures ranging from near zero kelvin to a few hundred degrees Celsius[4].
The STM is based on the concept of quantum tunnelling. When a conducting tip is brought very near to a metallic or semiconducting surface, a bias between the two can allow electrons to tunnel through the vacuum between them. For low voltages, this tunneling current is a function of the local density of states (LDOS) at the Fermi level, Ef, of the sample[4]. Variations in current as the probe passes over the surface are translated into an image. STM can be a challenging technique, as it requires extremely clean surfaces and sharp tips.
Contents |
Tunnelling is a functioning concept that arises from quantum mechanics. Classically, an object hitting an impenetrable barrier will not pass through. In contrast, objects with a very small mass, such as the electron, have wavelike characteristics which permit such an event, referred to as tunnelling.
Electrons behave as beams of energy, and in the presence of a potential U(z), assuming 1-dimensional case, the energy levels ψn(z) of the electrons are given by solutions to Schrödinger’s equation,
,where ħ is the reduced Planck’s constant, z is the position, and m is the mass of an electron[4]. If an electron of energy E is incident upon an energy barrier of height U(z), the electron wave function is a traveling wave solution,
,where

if E > U(z), which is true for a wave function inside the tip or inside the sample[4]. Inside a barrier, such as between tip and sample, E < U(z) so the wave functions which satisfy this are decaying waves,
,where

quantifies the decay of the wave inside the barrier, with the barrier in the +z direction for − κ [4].
Knowing the wave function allows one to calculate the probability density for that electron to be found at some location. In the case of tunneling, the tip and sample wave functions overlap such that when under a bias, there is some finite probability to find the electron in the barrier region and even on the other side of the barrier[4]. Let us assume the bias is V and the barrier width is W, as illustrated in Figure 1. This probability, P, that an electron at z=0 (left edge of barrier) can be found at z=W (right edge of barrier) is proportional to the wave function squared,
[4].If the bias is small, we can let U − E ≈ φM in the expression for κ, where φM, the work function, gives the minimum energy needed to bring an electron from an occupied level, the highest of which is at the Fermi level (for metals at T=0 kelvins), to vacuum level. When a small bias V is applied to the system, only electronic states very near the Fermi level, within eV, are excited[4]. These excited electrons can tunnel across the barrier. In other words, tunneling occurs mainly with electrons of energies near the Fermi level.
However, tunneling does require that there is an empty level of the same energy as the electron for the electron to tunnel into on the other side of the barrier. It is because of this restriction that the tunneling current can be related to the density of available or filled states in the sample. The current due to an applied voltage V (assume tunneling occurs sample to tip) depends on two factors: 1) the number of electrons between Ef and eV in the sample, and 2) the number among them which have corresponding free states to tunnel into on the other side of the barrier at the tip[4]. The higher density of available states the greater the tunneling current. When V is positive, electrons in the tip tunnel into empty states in the sample; for a negative bias, electrons tunnel out of occupied states in the sample into the tip[4].
Mathematically, this tunneling current is given by
.One can sum the probability over energies between Ef − eV and eV to get the number of states available in this energy range per unit volume, thereby finding the local density of states (LDOS) near the Fermi level[4]. The LDOS near some energy E in an interval ε is given by
,and the tunnel current at a small bias V is proportional to the LDOS near the Fermi level, which gives important information about the sample[4]. It is desirable to use LDOS to express the current because this value does not change as the volume changes, while probability density does[4]. Thus the tunneling current is given by

where ρs(0,Ef) is the LDOS near the Fermi level of the sample at the sample surface[4]. By using equation (6), this current can also be expressed in terms of the LDOS near the Fermi level of the sample at the tip surface,

The exponential term in (9) is very significant in that small variations in W greatly influence the tunnel current. If the separation is decreased by 1 Ǻ, the current increases by an order of magnitude, and vice versa[5].
This approach fails to account for the rate at which electrons can pass the barrier. This rate should affect the tunnel current, so it can be accounted for by using Fermi’s Golden Rule with the appropriate tunneling matrix element. John Bardeen solved this problem in his study of the metal-insulator-metal junction, MIM[6]. He found that if he solved Schrödinger’s equation for each side of the junction separately to obtain the wave functions ψ and χ for each electrode, he could obtain the tunnel matrix, M, from the overlap of these two wave functions[4]. This can be applied to STM by making the electrodes the tip and sample, assigning ψ and χ as sample and tip wave functions, respectively, and evaluating M at some surface S between the metal electrodes at z=zo, where z=0 at the sample surface and z=W at the tip surface[4].
Now, Fermi’s Golden Rule gives the rate for electron transfer across the barrier, and is written
,where δ(Eψ-Eχ) restricts tunneling to occur only between electron levels with the same energy[4]. The tunnel matrix element, given by
,is a description of the lower energy associated with the interaction of wave functions at the overlap, also called the resonance energy[4].
Summing over all the states gives the tunneling current as
,where f is the Fermi function, ρs and ρT are the density of states in the sample and tip, respectively[4]. The Fermi distribution function describes the filling of electron levels at a given temperature T.
First the tip is brought into close proximity of the sample by some coarse sample-to-tip control. The values for common sample-to-tip distance, W, range from about 4-7 Å, which is the equilibrium position between attractive (3<W<10Å) and repulsive (W<3Å) interactions[4]. Once tunneling is established, piezoelectric transducers are implemented to move the tip in three directions. As the tip is rastered across the sample in the x-y plane, the density of states and therefore the tunnel current changes. This change in current with respect to position can be measured itself, or the height, z, of the tip corresponding to a constant current can be measured[4]. These two modes are called constant height mode and constant current mode, respectively.
In constant current mode, feedback electronics adjust the height by a voltage to the piezoelectric height control mechanism[7]. This leads to a height variation and thus the image comes from the tip topography across the sample and gives a constant charge density surface; this means contrast on the image is due to variations in charge density[5].
In constant height, the voltage and height are both held constant while the current changes to keep the voltage from changing; this leads to an image made of current changes over the surface, which can be related to charge density[5]. The benefit to using a constant height mode is that it is faster, as the piezoelectric movements require more time to register the change in constant current mode than the voltage response in constant height mode[5].
In addition to scanning across the sample, information on the electronic structure of the sample can be obtained by sweeping voltage and measuring current at a specific location[3]. This type of measurement is called scanning tunneling spectroscopy (STS).
Framerates of at least 1 Hz enable so called Video-STM (up to 50 Hz possible). This can be used to scan surface diffusion.
The components of an STM include scanning tip, piezoelectric controlled height and x,y scanner, coarse sample-to-tip control, vibration isolation system, and computer[7].
The resolution of an image is limited by the radius of curvature of the scanning tip of the STM. Additionally, image artifacts can occur if the tip has two tips at the end rather than a single atom; this leads to “double-tip imaging,” a situation in which both tips contribute to the tunneling[3]. Therefore it has been essential to develop processes for consistently obtaining sharp, usable tips. Recently, carbon nanotubes have been used in this instance.
The tip is often made of tungsten or platinum-iridium, though gold is also used[3]. Tungsten tips are usually made by electrochemical etching, and platinum-iridium tips by mechanical shearing[3]. Both processes are outlined in C. Bai’s book, reference[3] below.
Due to the extreme sensitivity of tunnel current to height, proper vibration isolation is imperative for obtaining usable results. In the first STM by Binnig and Rohrer, magnetic levitation was used to keep the STM free from vibrations; now spring systems are often used[4]. Additionally, mechanisms for reducing eddy currents are implemented.
Maintaining the tip position with respect to the sample, scanning the sample in raster fashion and acquiring the data is computer controlled[7]. The computer is also used for enhancing the image with the help of image processing as well as performing quantitative morphological measurements.
Many other microscopy techniques have been developed based upon STM. These include Photon Scanning Microscopy (PSTM), which uses an optical tip to tunnel photons[3]; Scanning Tunneling Potentiometry (STP), which measures electric potential across a surface[3]; and spin polarized scanning tunneling microscopy (SPSTM), which uses a ferromagnetic tip to tunnel spin-polarized electrons into a magnetic sample[8].
Other STM methods involve manipulating the tip in order to change the topography of the sample. This is attractive for several reasons. Firstly the STM has an atomically precise positioning system which allows very accurate atomic scale manipulation. Furthermore, after the surface is modified by the tip, it is a simple matter to then image with the same tip, without changing the instrument. IBM researchers developed a way to manipulate Xenon atoms absorbed on a nickel surface[3] This technique has been used to create electron "corrals" with a small number of adsorbed atoms, which allows the STM to be used to observe electron Friedel Oscillations on the surface of the material. Aside from modifying the actual sample surface, one can also use the STM to tunnel electrons into a layer of E-Beam photoresist on a sample, in order to do lithography. This has the advantage of offering more control of the exposure than traditional Electron beam lithography. Another practical application of STM is atomic deposition of metals (Au, Ag, W, etc.) with any desired (pre-programmed) pattern, which can be used as contacts to nanodevices or as nanodevices themselves.
Recently groups have found they can use the STM tip to rotate individual bonds within single molecules. The electrical resistance of the molecule depends on the orientation of the bond, so the molecule effectively becomes a molecular switch.
An earlier, similar invention, the Topografiner of R. Young, J. Ward, and F. Scire from the NIST [9], relied on field emission. However, Young is credited by the Nobel Committee as the person who realized that it should be possible to achieve better resolution by using the tunnel effect [10].
| Wikibooks has a book on the topic of |
| Wikimedia Commons has media related to: Scanning tunneling microscope |
|
|||||||||||||||||
|
||||||||||||||||
This entry is from Wikipedia, the leading user-contributed encyclopedia. It may not have been reviewed by professional editors (see full disclaimer)
| Gerd Binnig (German physicist) | |
| Heinrich Rohrer (Swiss physicist) | |
| Ernst Ruska (German electrician & engineer) |
Copyrights:
![]() | Dictionary. The American Heritage® Dictionary of the English Language, Fourth Edition Copyright © 2007, 2000 by Houghton Mifflin Company. Updated in 2007. Published by Houghton Mifflin Company. All rights reserved. Read more | |
![]() | Sci-Tech Encyclopedia. McGraw-Hill Encyclopedia of Science and Technology. Copyright © 2005 by The McGraw-Hill Companies, Inc. All rights reserved. Read more | |
![]() | Columbia Encyclopedia. The Columbia Electronic Encyclopedia, Sixth Edition Copyright © 2003, Columbia University Press. Licensed from Columbia University Press. All rights reserved. www.cc.columbia.edu/cu/cup/ Read more | |
![]() | Wikipedia. This article is licensed under the GNU Free Documentation License. It uses material from the Wikipedia article "Scanning tunneling microscope". Read more |
Mentioned in