A substance that glows when hit by high-energy particles or photons.
| Wikipedia: Scintillator |
A scintillator is material which exhibits the property of luminescence[1] when excited by ionizing radiation. Luminescent materials, when struck by an incoming particle, absorb its energy and scintillate, i.e. reemit the absorbed energy in the form of a small flash of light, typically in the visible range. (Throughout this article, the word “particle” will be used to mean “ionizing radiation” and can refer to either charged particulate radiation such as electrons and heavy charged particles, or to uncharged radiation such as photons and neutrons, provided that they have enough energy to induce ionization.) If the reemission occurs promptly, i.e. within the ≈10−8s required for an atomic transition, the process is called fluorescence. Sometimes, the excited state is metastable, so the relaxation back out of the excited state is delayed (necessitating anywhere from a few microseconds to hours depending on the material): the process then corresponds to either one of two phenomena, depending on the type of transition and hence the wavelength of the emitted optical photon: delayed fluorescence or phosphorescence (also called after-glow).
A scintillation detector or scintillation counter is obtained when a scintillator is coupled to an electronic light sensor such as a photomultiplier tube (PMT) or a photodiode. PMTs absorb the light emitted by the scintillator and reemit it in the form of electrons via the photoelectric effect. The subsequent multiplication of those electrons (sometimes called photo-electrons) results in an electrical pulse which can then be analyzed and yield meaningful information about the particle that originally struck the scintillator. Vacuum photodiodes are similar but do not amplify the signal while silicon photodiodes accomplish the same thing directly in the silicon.
The first use of a scintillator dates back to an experiment in 1903 where Sir William Crookes observed a ZnS screen struck by α-particles.[citation needed] The scintillations produced by the screen were visible to the naked eye if viewed by a microscope in a darkened room; the device was known as a spinthariscope. The technique led to a number of important discoveries but was obviously tedious. Scintillators gained additional attention in 1944, when Curran and Baker replaced the naked eye measurement with the newly developed PMT. This was the birth of the modern scintillation detector.
Today, scintillation detectors are used in a wide array of applications, including fundamental research in particle and nuclear physics, oil exploration, field spectrometry, container and baggage scanning, health physics, space physics, industrial gauging, and medical diagnostics and therapy (PET, SPECT, therapy imaging, etc).
Contents |
Among the properties desirable in a good detector scintillator are: a low gamma output (i.e. a high efficiency for converting the energy of incident radiation into scintillation photons), transparency to its own scintillation light (for good light collection), efficient detection of the radiation being studied, a high stopping power, good linearity over a wide range of energy, a short rise time for fast timing applications (e.g. coincidence measurements), a short decay time to reduce detector dead-time and accommodate high event rates, emission in a spectral range matching the spectral sensitivity of existing PMTs (although wavelength shifters can sometimes be used), an index of refraction near that of glass (≈1.5) to allow optimum coupling to the PMT window. Ruggedness and good behavior under high temperature may be desirable where resistance to vibration and high temperature is necessary (e.g. oil exploration). The practical choice of a scintillator material is usually a compromise between those properties to best fit a given application.
Among the properties listed above, the light output is the most important, as it affects both the efficiency and the resolution of the detector (the efficiency is the ratio of detected particles to the total number of particles impinging upon the detector; the energy resolution is the ratio of the full width at half maximum of a given energy peak to the peak position, usually expressed in %). The light output is a strong function of the type of incident particle or photon and of its energy, which therefore strongly influences the type of scintillation material to be used for a particular application. The presence of quenching effects results in reduced light output (i.e. reduced scintillation efficiency). Quenching refers to all radiationless deexcitation processes in which the excitation is degraded mainly to heat.[2] The overall signal production efficiency of the detector, however, also depends on the quantum efficiency of the PMT (typically ≈30% at peak), and on the efficiency of light transmission and collection (which depends on the type of reflector material covering the scintillator and light guides, the length/shape of the light guides, any light absorption, etc). The light output is often quantified as a number of scintillation photons produced per keV of deposited energy. Typical numbers are (when the incident particle is an electron): ≈40 photons/keV for NaI(Tl), ≈10 photons/keV for plastic scintillators, and ≈4 photons/keV for BGO.
Scintillation detectors are generally assumed to be linear. This assumption is based on two requirements: (1) that the light output of the scintillator is proportional to the energy of the incident radiation; (2) that the electrical pulse produced by the photomultiplier tube is proportional to the emitted scintillation light. The linearity assumption is usually a good rough approximation, although deviations can occur (especially pronounced for particles heavier than the proton at low energies).
Resistance and good behavior under high-temperature, high-vibration environments is especially important for applications such as oil exploration (wireline logging, measurement while drilling). For most scintillators, light output depends on the temperature. This dependence can largely be ignored for room-temperature applications since it is usually weak. The dependence on the temperature is also weaker for organic scintillators than it is for inorganic crystals, such as ZnS(Ag) or BGO. The coupled PMTs also exhibit temperature sensitivity, and can be damaged if submitted to mechanical shock. Hence, high temperature rugged PMTs should be used for high-temperature, high-vibration applications.
The time evolution of the number of emitted scintillation photons N in a single scintillation event can often be described by the linear superposition of one or two exponential decays. For two decays, we have the form:

where τf and τs are the fast (or prompt) and the slow (or delayed) decay constants. Many scintillators are characterized by 2 time components: one fast (or prompt), the other slow (or delayed). While the fast component usually dominates, the relative amplitude A and B of the two components depend on the scintillating material. Both of these components can also be a function the energy loss dE/dx. In cases where this energy loss dependence is strong, the overall decay time constant varies with the type of incident particle. Such scintillators enable pulse shape discrimination, i.e. particle identification based on the decay characteristics of the PMT electric pulse. For instance, when BaF2 is used, γ rays typically excite the fast component, while α particles excite the slow component: it is thus possible to identify them based on the decay time of the PMT signal.
Organic scintillators are aromatic hydrocarbon compounds containing linked or condensed benzene ring structures.[1] They typically have a very rapid decay time (a few ns or less).
Some organic scintillators are pure crystals. The most common types are anthracene (C14H10, decay time ≈30 ns), stilbene (C14H12, few ns decay time), and naphthalene (C10H8, few ns decay time). They are very durable, but their response is anisotropic (which spoils energy resolution when the source is not collimated), and they cannot be easily machined, nor can they be grown in large sizes; hence they are not very often used. Anthracene has the highest light output of all organic scintillators and is therefore chosen as a reference: the light outputs of other scintillators are sometimes expressed as a percent of anthracene light.
These are liquid solutions of one or more organic scintillators in an organic solvent. The typical solutes are fluors such as p-Terphenyl (C18H14), PBD (C20H14N2O), butyl PBD (C24H22N2O), PPO (C15H11NO), and wavelength shifter such as POPOP (C24H16N2O). The most widely used solvents are toluene, xylene, benzene, phenylcyclohexane, triethylbenzene, and decalin. Liquid scintillators are easily loaded with other additives such as wavelength shifters to match the spectral sensitivity range of a particular PMT, or 10B to increase the neutron detection efficiency of the scintillation counter itself (since 10B has a high interaction cross section with thermal neutrons). For many liquids, dissolved oxygen can act as a quenching agent and lead to reduced light output, hence the necessity to seal the solution in an oxygen-free, air-tight enclosure.[2]
Plastic scintillators are solutions of organic scintillators in a solvent which is subsequently polymerized to form a solid. Some of the common solutes are p-Terphenyl, PBD, b-PBD, PBO, POPOP. The most widely used plastic solvents are polyvinyltoluene and polystyrene.
Plastics scintillators give a fast signal (a few ns) and a high light output. The number of emitted scintillation photons is best described by the convolution of an exponential decay and a Gaussian (rather than the exponential decay alone):[1]

where the function f is a Gaussian.
Plastics can very easily be shaped and machined to the forms desired in detectors (cylinders, rods, flat sheets, fibers, and thin films) and are relatively inexpensive, which is why they are among the most widely used scintillators. They are generally quite resistant but can be scratched and are easily attacked by organic solvents (e.g. acetone). Also, bodily acids from one’s hands can cause cracking over time, so handling should be done using cotton or terylene gloves.[1]
Inorganic scintillators are usually crystals grown in high temperature furnaces, for example, alkali metal halides, often with a small amount of activator impurity. The most widely used is NaI(Tl) (sodium iodide doped with thallium). Other inorganic alkali halide crystals are: CsI(Tl), CsI(Na), CsI(pure), CsF, KI(Tl), LiI(Eu). Some non-alkali crystals include: BaF2, CaF2(Eu), ZnS(Ag), CaWO4, CdWO4, YAG(Ce) (Y3Al5O12(Ce)), GSO, LSO. (For more examples, see also phosphors.
Newly developed products include LaCl3(Ce), lanthanum chloride doped with Cerium, as well as a Cerium-doped lanthanum bromide, LaBr3(Ce). They are both very hygroscopic (i.e., damaged when exposed to moisture in the air) but offer excellent light output and energy resolution (63 photons/keV γ for LaBr3(Ce) versus 38 photons/keV γ for NaI(Tl)), a fast response (16 ns for LaBr3(Ce) versus 250 ns for NaI(Tl)), excellent linearity, and a very stable light output over a wide range of temperatures. In addition LaBr3(Ce) offers a higher stopping power for γ rays (density of 5.08 g/cm3 versus 3.67 g/cm3 for NaI(Tl)). LYSO (Lu1.8Y0.2SiO5(Ce)) has an even higher density (7.1 g/cm3, comparable to BGO), is non-hygroscopic, and has a higher light output than BGO (32 photons/keV γ), in addition to being rather fast (41 ns decay time versus 300 ns for BGO).
A disadvantage of some inorganic crystals, e.g., NaI, is their hygroscopicity, a property which requires them to be housed in an air-tight enclosure to protect them from moisture. CsI(Tl) and BaF2 are only slightly hygroscopic and do not usually need protection. CsF, NaI(Tl), LaCl3(Ce), LaBr3(Ce) are hygroscopic, while BGO, CaF2(Eu), LYSO, and YAG(Ce) are not.
Inorganic crystals can be cut to small sizes and arranged in an array configuration so as to provide position sensitivity. Such arrays are often used in medical physics or security applications to detect X-rays or γ rays: high-Z, high density materials (e.g., LYSO, BGO) are typically preferred for this type of applications.
Scintillation in inorganic crystals is typically slower than in organic ones, ranging typically from 250 ns for NaI(Tl) to 1000 ns for CsI(Tl). Exceptions are CsF (≈5 ns), fast BaF2 (0.7 ns; the slow component is at 630 ns), as well as the newer products (LaCl3(Ce), 28 ns; LaBr3(Ce), 16 ns; LYSO, 41 ns).
Gaseous scintillators consist of nitrogen and the noble gases helium, argon, krypton, and xenon, with helium and xenon receiving the most attention. The scintillation process is due to the de-excitation of single atoms excited by the passage of an incoming particle. This de-excitation is very rapid (≈1 ns), so the detector response is quite fast. Coating the walls of the container with a wavelength shifter is generally necessary as those gases typically emit in the ultraviolet and PMTs respond better to the visible blue-green region. In nuclear physics, gaseous detectors have been used to detect fission fragments or heavy charged particles.
The most common glass scintillators are cerium-activated lithium or boron silicates. Since both lithium and boron have large neutron cross-sections, glass detectors are particularly well suited to the detection of thermal (slow) neutrons. Lithium is more widely used than boron since it has a greater energy release on capturing a neutron and therefore greater light output. Glass scintillators are however sensitive to electrons and γ rays as well (pulse height discrimination can be used for particle identification). Being very robust, they are also well-suited to harsh environmental conditions. Their response time is ≈10 ns, their light output is however low, typically ≈30% of that of anthracene.
Transitions made by the free valence electrons of the molecules are responsible for the production of scintillation light in organic crystals.[1] These electrons are associated with the whole molecule rather than any particular atom and occupy the so-called -molecular orbitals. The ground state S0 is a singlet state above which are the excited singlet states (S*, S**,…), the lowest triplet state (T0), and its excited levels (T*, T**,…). A fine structure corresponding to molecular vibrational modes is associated with each of those electron levels. The energy spacing between electron levels is ≈1 eV; the spacing between the vibrational levels is about 1/10th of that for electron levels.
An incoming particle can excite either an electron level or a vibrational level. The singlet excitations immediately decay (< 10 ps) to the S* state without the emission of radiation (internal degradation). The S* state then decays to the ground state S0 (typically to one of the vibrational levels above S0) by emitting a scintillation photon. This is the prompt component or fluorescence. The transparency of the scintillator to the emitted photon is due to the fact that the energy of the photon is less than that required for a S* → S0 transition (the transition is usually being to a vibrational level above S0).
When one of triplet states gets excited, it immediately decays to the T0 state with no emission of radiation (internal degradation). Since the T0 → S0 transition is very improbable, the T0 state instead decays by interacting with another T0 molecule:

and leaves one of the molecules in the S* state, which then decays to S0 with the release of a scintillation photon. Since the T0-T0 interaction takes time, the scintillation light is delayed: this is the slow or delayed component (corresponding to delayed fluorescence). Sometimes, a direct T0 → S0 transition occurs (also delayed), and corresponds to the phenomenon of phosphorescence (note that the difference between delayed-fluorescence and phosporescence lies in the difference in the wavelengths of the emitted optical photon in a S* → S0 transition versus a T0 → S0 transition).
Organic scintillators can be dissolved in an organic solvent to form either a liquid or plastic scintillator. The scintillation process is the same as described for organic crystals (above); what differs is the mechanism of energy absorption: energy is first absorbed by the solvent, then passed onto the scintillation solute (the details of the transfer are not clearly understood).
The scintillation process in inorganic materials is due to the electronic band structure found in crystals and is not molecular in nature as is the case with organic scintillators.[1] An incoming particle can excite an electron from the valence band to either the conduction band or the exciton band (located just below the conduction band and separated from the valence band by an energy gap; see picture). This leaves an associated hole behind, in the valence band. Impurities create electronic levels in the forbidden gap. The excitons are loosely bound electron-hole pairs which wander through the crystal lattice until they are captured as a whole by impurity centers. The latter then rapidly de-excite by emitting scintillation light (fast component). The activator impurities are typically chosen so that the emitted light is in the visible range or near UV where PMTs are effective. The holes associated with electrons in the conduction band are independent from the latter. Those holes and electrons are captured successively by impurity centers exciting certain metastable states not accessible to the excitons. The delayed de-excitation of those metastable impurity states again results in scintillation light (slow component).
BGO is a pure inorganic scintillator without any activator impurity. There, the scintillation process is due to an optical transition of the Bi3+ ion, a major constituent of the crystal.[2] A similar process exists in CdWO4.
In gases, the scintillation process is due to the de-excitation of single atoms excited by the passage of an incoming particle (a very rapid process: ≈1 ns).
Scintillation counters are usually not ideal for the detection of heavy ions for three reasons: (1) the very high ionizing power of heavy ions induces quenching effects which result in a reduced light output (e.g. for equal energies, a proton will produce ¼ to ½ the light of on electron, while alphas will produce only about 1/10th the light; cf SGC website); (2) the high dE/dx also results in a reduction of the fast component relative to the slow component, increasing detector dead-time; (3) strong non-linearities are observed in the detector response especially at lower energies. The reduction in light output is stronger for organics than for inorganic crystals. Therefore, where needed, inorganic crystals, e.g. CsI(Tl), ZnS(Ag) (typically used in thin sheets as α-particle monitors) , CaF2(Eu), should be preferred to organic materials. Typical applications are α-survey instruments, dosimetry instruments, and heavy ion dE/dx detectors.[3] Gaseous scintillators have also been used in nuclear physics experiments.
The detection efficiency for electrons is essentially 100% for most scintillators. But, because electrons can make large angle scatterings (sometimes back-scatterings), they can exit the detector without depositing their full energy in it. The back-scattering is a rapidly increasing function of the atomic number Z of the scintillator material. Organic scintillators, having a lower Z than inorganic crystals, are therefore best suited for the detection of low-energy (< 10 MeV) electrons. The situation is different for high energy electrons: since they mostly lose their energy by bremsstrahlung at the higher energies, a higher-Z material is better suited for the detection of the bremsstrahlung photon and the production of the electromagnetic shower which it can induce.
High-Z materials, e.g. inorganic crystals, are best suited for the detection of gamma rays. The three basic ways that a gamma ray interacts with matter are: the photoelectric effect, Compton scattering, and pair production. The photon is completely absorbed in photoelectric effect and pair production, while only partial energy is deposited in any given Compton scattering. The cross section for the photoelectric process is proportional to Z5, that for pair production proportional to Z2, whereas Compton scattering goes roughly as Z. A high-Z material therefore favors the former two processes, enabling the detection of the full energy of the gamma ray.
Since the neutron is not charged it does not interact via the Coulomb force and therefore does not ionize the scintillation material. It must first transfer some or all of its energy via the strong force to a charged nucleus. The positively charged nucleus then produces ionization. Fast neutrons (generally >0.5 MeV [2]) primarily rely on the recoil proton in (n,p) reactions; materials rich in hydrogen, e.g. plastic scintillators, are therefore best suited for their detection. Slow neutrons rely on nuclear reactions such as the (n,γ) or (n,α) reactions, to produce ionization. Their mean free path is therefore quite large unless the scintillator material is loaded with elements having a high cross section for these nuclear reactions such as 6Li or 10B. Materials such as LiI(Eu) or glass silicates are therefore particularly good for the detection of slow (thermal) neutrons.
The following is a list of commonly used inorganic crystals:
This entry is from Wikipedia, the leading user-contributed encyclopedia. It may not have been reviewed by professional editors (see full disclaimer)
| wavelength shifter (electronics) | |
| photon emission spectrum (physics) | |
| Scintillation counter (nuclear physics) |
| How is a scintillation form in the crystal? Read answer... | |
| What type of radiation is best detected by a scintillation counter? Read answer... | |
| How do you spell scintillating? Read answer... |
| What is a Frequency Scintillation? | |
| What is scintillation counter? | |
| What Scintillate mean? |
Copyrights:
![]() | Dictionary. The American Heritage® Dictionary of the English Language, Fourth Edition Copyright © 2007, 2000 by Houghton Mifflin Company. Updated in 2009. Published by Houghton Mifflin Company. All rights reserved. Read more | |
![]() | Medical Dictionary. The American Heritage® Stedman's Medical Dictionary Copyright © 2002, 2001, 1995 by Houghton Mifflin Company. Read more | |
![]() | Wikipedia. This article is licensed under the Creative Commons Attribution/Share-Alike License. It uses material from the Wikipedia article "Scintillator". Read more |
Mentioned in